Nissei DC-100 and Qsafe QSF 104 pocket Szintillator
Inviato: 11/03/2016, 19:46
Hello,
here my Qsafe and Nissei pocket scintillators.
Nissei Radcounter DC-100 with CsI (TI) weight 62 gramm
Qsafe QSF 104 scintillator CsI (TI) weight 140 gramm
The Qsafe have in 1h 15000 counts. See picture.
My background is 0,13 uSv/h
The Qsafe have a gamma sensitivity over 2000cpm/uSv/h
The gamma sensitivity from DC-100 is unknown.
No technical data.
I read a background with Nissei DC-100 0.05uSv/h - 0,07uSv/h
The Qsafe QSF104 read 0,12uSv/h - 0,13 uSv/h. and 260 cpm
Background:
DC-100 0,05-0,07uSv/h
Qsafe Qsf104 0,12 usv/h and 256cpm
Test with a AM241 pill from a smoke detector protected in a thin plastic housing.
DC-100 0,07uSv/h - 0,09uSv/h have not display for read cpm, not sensitive Am241 gamma ??
Qsafe Qsf104 3,80 usv/h and 8150cpm
Test with LU176
DC-100 0,06 uSv/h - 0,07 uSv/h
Qsafe Qsf104 0,46 uSv/h 980cpm
Test with CS137 spark gap tube
DC-100 0,42 uSv/h device is working. very high response, with a counter with SBM20 tube i read 0,27 - 0,32uSv/h !
Qsafe Qsf104 0,28 uSv/h 615cpm
Test with a gas mantle Thorium
DC-100 0,88 uSv/h , with a counter with SBM20 tube i read 0,64uSv/h with gamma filter
Qsafe Qsf104 0,73 uSv/h 1580cpm
The Qsafe range: 50kev - 6Mev
here intern pictures Nissei DC-100
Thanks for watching!
here my Qsafe and Nissei pocket scintillators.
Nissei Radcounter DC-100 with CsI (TI) weight 62 gramm
Qsafe QSF 104 scintillator CsI (TI) weight 140 gramm
The Qsafe have in 1h 15000 counts. See picture.
My background is 0,13 uSv/h
The Qsafe have a gamma sensitivity over 2000cpm/uSv/h
The gamma sensitivity from DC-100 is unknown.
No technical data.
I read a background with Nissei DC-100 0.05uSv/h - 0,07uSv/h
The Qsafe QSF104 read 0,12uSv/h - 0,13 uSv/h. and 260 cpm
Background:
DC-100 0,05-0,07uSv/h
Qsafe Qsf104 0,12 usv/h and 256cpm
Test with a AM241 pill from a smoke detector protected in a thin plastic housing.
DC-100 0,07uSv/h - 0,09uSv/h have not display for read cpm, not sensitive Am241 gamma ??
Qsafe Qsf104 3,80 usv/h and 8150cpm
Test with LU176
DC-100 0,06 uSv/h - 0,07 uSv/h
Qsafe Qsf104 0,46 uSv/h 980cpm
Test with CS137 spark gap tube
DC-100 0,42 uSv/h device is working. very high response, with a counter with SBM20 tube i read 0,27 - 0,32uSv/h !
Qsafe Qsf104 0,28 uSv/h 615cpm
Test with a gas mantle Thorium
DC-100 0,88 uSv/h , with a counter with SBM20 tube i read 0,64uSv/h with gamma filter
Qsafe Qsf104 0,73 uSv/h 1580cpm
The Qsafe range: 50kev - 6Mev
here intern pictures Nissei DC-100
Thanks for watching!