Hello,
here my Qsafe and Nissei pocket scintillators.
Nissei Radcounter DC-100 with CsI (TI) weight 62 gramm
Qsafe QSF 104 scintillator CsI (TI) weight 140 gramm
The Qsafe have in 1h 15000 counts. See picture.
My background is 0,13 uSv/h
The Qsafe have a gamma sensitivity over 2000cpm/uSv/h
The gamma sensitivity from DC-100 is unknown.
No technical data.
I read a background with Nissei DC-100 0.05uSv/h - 0,07uSv/h
The Qsafe QSF104 read 0,12uSv/h - 0,13 uSv/h. and 260 cpm
Background:
DC-100 0,05-0,07uSv/h
Qsafe Qsf104 0,12 usv/h and 256cpm
Test with a AM241 pill from a smoke detector protected in a thin plastic housing.
DC-100 0,07uSv/h - 0,09uSv/h have not display for read cpm, not sensitive Am241 gamma ??
Qsafe Qsf104 3,80 usv/h and 8150cpm
Test with LU176
DC-100 0,06 uSv/h - 0,07 uSv/h
Qsafe Qsf104 0,46 uSv/h 980cpm
Test with CS137 spark gap tube
DC-100 0,42 uSv/h device is working. very high response, with a counter with SBM20 tube i read 0,27 - 0,32uSv/h !
Qsafe Qsf104 0,28 uSv/h 615cpm
Test with a gas mantle Thorium
DC-100 0,88 uSv/h , with a counter with SBM20 tube i read 0,64uSv/h with gamma filter
Qsafe Qsf104 0,73 uSv/h 1580cpm
The Qsafe range: 50kev - 6Mev
here intern pictures Nissei DC-100
Thanks for watching!
Nissei DC-100 and Qsafe QSF 104 pocket Szintillator
Nissei DC-100 and Qsafe QSF 104 pocket Szintillator
Non hai i permessi necessari per visualizzare i file allegati in questo messaggio.
Ultima modifica di pg103 il 11/03/2016, 20:23, modificato 1 volta in totale.
- marconmeteo
- Moderatore
- Messaggi: 4392
- Iscritto il: 08/10/2015, 21:20
- Località: Marcon (VE)
- Contatta:
Re: Nissei DC-100 and Qsafe QSF 104 pocket Szintillator
Thank you, pg103, for this nice roundup of instruments, also related to technical specifications and tests. Thank you, because I did not know these tools.
[External Link Removed for Guests]
"Due cose sono infinite: l'universo e la stupidità umana, ma riguardo all'universo ho ancora dei dubbi" - Albert Einstein
IU3ARP
"Due cose sono infinite: l'universo e la stupidità umana, ma riguardo all'universo ho ancora dei dubbi" - Albert Einstein
IU3ARP
Re: Nissei DC-100 and Qsafe QSF 104 pocket Szintillator
Really very interesting!
Thanks
Thanks
[External Link Removed for Guests]
Re: Nissei DC-100 and Qsafe QSF 104 pocket Szintillator
Thanks for the answers
The QSF-104 model comes from korea.
The DC-100 from Japan.
The sensor of QSF104 is shock resistant.
The sensor of DC-100 is not shock resistant.
If I knock more with the finger on the CsI (TI) sensor of dc-100,
then the value is clearly upwards.
QSF104 have not this problem.
According to the manufacturer, it is possible to detect 100 BQ CS137.
Here the interesting manual for qsf104 with measurement technical details for this device.
Is interesting to see.
There is a new model that QSF104XE.
It has an external larger CsI (TI) unit.
See photo:
The QSF104XE have a gamma sensitivity : 6000cpm/uSv/h !
My modell: 2000cpm/uSv/h.
Ciao
The QSF-104 model comes from korea.
The DC-100 from Japan.
The sensor of QSF104 is shock resistant.
The sensor of DC-100 is not shock resistant.
If I knock more with the finger on the CsI (TI) sensor of dc-100,
then the value is clearly upwards.
QSF104 have not this problem.
According to the manufacturer, it is possible to detect 100 BQ CS137.
Here the interesting manual for qsf104 with measurement technical details for this device.
Is interesting to see.
There is a new model that QSF104XE.
It has an external larger CsI (TI) unit.
See photo:
The QSF104XE have a gamma sensitivity : 6000cpm/uSv/h !
My modell: 2000cpm/uSv/h.
Ciao
Non hai i permessi necessari per visualizzare i file allegati in questo messaggio.
Re: Nissei DC-100 and Qsafe QSF 104 pocket Szintillator
Very interesting, thank for your photos.
I've never seen them.
I've never seen them.
I nostri successi e i nostri fallimenti sono tra loro inscindibili, proprio come la materia e l'energia. Se vengono separati, l'uomo muore. Nikola Tesla